HP/83624A

Started by dminesinger, 05-29-2009 -- 02:26:20

Previous topic - Next topic

dminesinger

Have an HP/83624A that when pinged on the IEEE488 returns with an ID on HP/83620A. When cal'd 9 months ago it worked fine. Any ideas?

PapaBear
PapaBear

*'It's the Veteran, not the reporter who has given us the freedom of the press.'
'It's the Veteran, not the poet, who has given us the freedom of speech.'
''It's the Military who salutes the flag, who serves beneath the flag, and whose coffin is draped by the flag, who allows the protester to burn the flag.'

flow

Check the unit's Identifier in the Calibration Constants section of the service manual! That should solve your problem.

dminesinger

Flow, Thank you. That solved the problem. Think the battery is going bad and not holding the calibration constants.

PB


Quote from: flow on 06-01-2009 -- 16:52:25
Check the unit's Identifier in the Calibration Constants section of the service manual! That should solve your problem.
PapaBear

*'It's the Veteran, not the reporter who has given us the freedom of the press.'
'It's the Veteran, not the poet, who has given us the freedom of speech.'
''It's the Military who salutes the flag, who serves beneath the flag, and whose coffin is draped by the flag, who allows the protester to burn the flag.'

MRD

PB,

  In my experience, I've seen the 8360s constants being changed to reflect others in the series, sometimes to accommodate old automated software that engineers had custom made and also to change for calibration, temporarily.  Once upon a time, before SureCal came out with 83640A/B/L and 83650A/B/L procedures we'd change them to an 83630A/B/L and do the rest manually. 

Mark

CalLabSolutions

That is why we stopped selling procedure based on the model number.  People always find away around your security, and adding in all that security makes the procedures harder to support.

Now we sell one procedure that covers ever model number in the manual.  And we are able to price our 836xx Series procedure under SureCal's price for the whole series and include all the optional test points.



 
Michael L. Schwartz