When is Data Required?

Started by briansalomon, 06-22-2017 -- 14:52:12

Previous topic - Next topic

briansalomon

I am working in an accredited (PJLA) lab and am not crystal clear on where it is defined which standards need data w/uncertainty and which do not.

We have been operating on the premise that primary standards require it but secondary standards do not.

I have been studying the PJLA Measurement Traceability Policy and don't see a clear definition there.

Bring technical excellence with you when you walk in the door every day.

Snap

If you are an accredited calibration lab, your standards all need data and uncertainties or else you cannot calculate uncertainties for your customers. 

briansalomon

Well I'm here to learn. You're right.

PL-2 says in 2.7.3   "a stated measurement result or value, with a
documented uncertainty" has to be included even for in-house calibartations.

I don't mind a little egg on my face as long as I learn something.

Thanks.
Bring technical excellence with you when you walk in the door every day.

dminesinger

We should learn something new every day. When I stop learning. I'll be 6 ft under or in a coma.