Fluke 5520A error 509 output exceeds user limits.

Started by USMCPMEL, 09-23-2010 -- 22:11:39

Previous topic - Next topic

USMCPMEL

This error comes up when i try to output above 11 amps on my Fluke 5520A any idea why?

USMCPMEL

LOL ok just fooled around with it and got it. After it came back from calibration someone must have altered one of the settings on the instrument. I pressed the setup kep then instmt setup then output setup then set limits then current then upper limit then I inputed 20.5 Amps for the upper limit and saved it.

mdbuike

Ya know there's a reason it goes to 20 amps, don't ya?

So airmen can check out 10 amp fuses   :-D   