BRUNSON 272-11 Multitarget Collimator Cal Procedure

Started by ZZ, 05-21-2020 -- 11:25:23

Previous topic - Next topic

ZZ

Just wondering if anyone knows which cal procedure (33K/Navair/TM) is for the Brunson 272-11 multitarget collimator? Searched GIDEP but only get procedures for alignment collimators. Thanks

GA_PMEL_RF

#1
May be 272-15 procedure may be the same procedure. I cal'd our 272-15 at Robins and used a ccd camera attachment with laptop to view the scope to save on neck and eye strain.
Works well.  Skip to the end of the images gallery
Skip to the beginning of the images gallery
AlignCam® System (Brunson)

GA_PMEL_RF

#2
Try Procedure 33k6-4-926-1.  ALIGNMENT COLLIMATOR GENERAL  It's on GIDEP
Covers Multitarget collimators. 