Equipment Substitution for Diff Voltmeter

Started by griff61, 11-08-2006 -- 09:14:52

Previous topic - Next topic

griff61

:? I'm a bit hazy on this one, can anyone help me with what can be substituted for a Differential Voltmeter? I have a single item that calls for one. Don't see buying one any time soon, but I'd sure like to avoid sending it elsewhere.
Sarcasm - Just one more service I offer

K-Rock

Depends on the application, but if loading is not an issue for your application, then a good bench multimeter should serve your purpose...

Hoopty

I used to use the 3458 in place of the diff voltmeter alot, but k-rock is right, check the impedance requirements.  The extremely high (GOhm?) input impedance was the main reason for using one.
#FDJT

Hoopty

Just looked up the specs on 3458.  The input impedance for 10V range and below is >10GOhm.  100V & 1000V ranges are 10 MOhm.  So it should be a suitable in any case as long as you use the right range.
#FDJT

±0.03 PPM Drifter

#4
You can use a null detector with a separate DC Standard if you want lots of High impedance. I use a Fluke 8508A Referance multimeter sometimes and an HP34401A on 10G input impedance, for 1V RNG or less, when I don't want to put un-needed usage on the 8508A. 3458A's are nice also. You also can put calibrated a 1G or 10G resistor in series with a 10M ohm DMM, taking into acount for the impedence of the DMM.