do you know vm2710a code CAL:SEC:STAT?. it is writed in user's manual, cal secure off code.
but, invalide code. error alram on.
do you know. what should i do?
나의 IM-100S 의 Tapatalk에서 보냄
CAL:SEC:STAT? is a query to tell you if cal mode is locked or unlocked.
This is a complicated procedure. The manual does not give good instructions, and if you are not careful, it is easy to incorrectly adjust the unit.
To unlock the calibration, Press CONFIGURE, CALIBRATION.
Click the CAL:SEC:STAT to turn it off, then enter password "VM2710" and click OK.
That will unlock the unit for calibration adjustment. And you need to set to appropriate range for each adjustment. After you do the adjustment you want, press CAL:STOR, the click CAL:SEC:STAT from in the calibration screen (not the main operating screen). Then press INSTRUMENT>RESET.
Again, this is not an intuitive adjustment. Take a lot of notes, and record all coefficient values from the calibration screen (they are different for each range).