When calibrating the 5720A is it necessary to characterize another 5720A and include it in the calibration to achieve 4:1 accuracy?
well, what you would do is to verify the output against a sufficiently accurate standard. that would not be a 5720a that has just been self-calibrated (artifact calibration). if that answers your question?
I got the 5790A, 8508A, DC references, resistors, AC shunts, reference divider, and null detector. I just wasn't sure if I also needed to use a calibrator in this process. Trying to figure out how they would do it at Fluke in Everett.