Looking for a calibration procedure for a Amsafe 508668-201 Diagnostic Tool
Did you contact the manufacturer? I did a quick check and was not able to find anything.
I know this post was a while ago... but did you ever have any luck finding a procedure?
Has anyone found any procedure for this? Would really like to be able to calibrate these in-house, currently have 50 of these units. Looks like some votage measurements and maybe light checks.
Did anyone find a procedure? I have looked and called the manufacturer multiple time with no resolution. I have a customer that needs it certified asap
Thanks!
There appears to be no written procedure for this.
I looked at this document https://customervault.amsafe.com/BinaryData/E508750G.pdf?binID=9084
There are general specs but no toleranced parameters given for this instrument. It appears to simply monitoring switch positions.
You can either send it back to the OEM or dig into the specifications of the equipment it is used to verify and write your own procedure based on them.