Looking for a 3rd Party calibration Supplier for this instrument. Range is 700 to 2000 Deg C. Typically calibrated with a blockbody source. Mikron (LumaSense) has a 7 to 9 week lead time.
I've checked with the usual suspects (Tektronix Service Solutions, ESI, Fluke (Hart), Simco) without any luck. Process Instruments can cal only to 1700 Deg C.
Any help appreciated...
Found that Fluke (Raytek) can perform these calibrations to 2700 Deg C.
Anyone have good thoughts on an accredited provider for one of those simple Hart sources (50/100/150 Deg C output blackbody reference)?
Not sure if they can cover the range you need but Bionetics/AFPSL does calibrate pyrometers and blackbodys. The contact point for commercial cals is Alice Machwart, Phone: (740) 788-5411.
Can't hurt to check.
John Treekiller
Good input. Thanks very much. I would think my range should be pretty easy (50 / 100 / 150 Deg C). I will give them a call.
Hart (Fluke) in Utah can probably do that for you.