Title: Fluke 5520A error 509 output exceeds user limits.
Post by: USMCPMEL on 09-23-2010 -- 22:11:39
This error comes up when i try to output above 11 amps on my Fluke 5520A any idea why?
Title: Re: Fluke 5520A error 509 output exceeds user limits.
Post by: USMCPMEL on 09-23-2010 -- 22:15:40
LOL ok just fooled around with it and got it. After it came back from calibration someone must have altered one of the settings on the instrument. I pressed the setup kep then instmt setup then output setup then set limits then current then upper limit then I inputed 20.5 Amps for the upper limit and saved it.
Title: Re: Fluke 5520A error 509 output exceeds user limits.
Post by: mdbuike on 09-24-2010 -- 02:49:29
Ya know there's a reason it goes to 20 amps, don't ya?

So airmen can check out 10 amp fuses   :-D   