I believe the readings are out of range and it will not take the new cal constants on 4 wire ohms. I have got everything else to work so far and it comes down to this. 1 k ohms is reading 1.00186 and the other ranges are sililar to this one. Any ideas?
What are you using for a standard? Just curious.
5520A with a fresh calibration. (smelling ) umm it has that just cal'd smell!!!
Well, we stand in the same boat, I'm doing a cal on one of those right now myself. The 5520 may not be accurate enough for the 4 wire ohms measurement, and definatly not for the rest of the readings. If you have a 5700 or equivilent thats your best bet. I'd recommend verifying the ohms out from the '20 with an accurate multimeter with 7 1/2 dig's of res. Thats the best I got for ya man.
What readings is it not accurate enough for??
Attached is a datasheet for a 34970A with TAR's calculated with the 5520A 1 year specs. Because of all the less than 4:1's we use a 5720A and 5790A. Hope this helps. Let me know if you need calculated TUR's. Additionally; I'm not a big fan of the 'synthesized' resistance (or whatever you call it in the 55X0A series), you may want to try a true resistance.