PMEL Forum

General => General Discussion => Topic started by: Duckbutta on 01-18-2009 -- 23:00:03

Title: Measurement Uncertainty is Overrated
Post by: Duckbutta on 01-18-2009 -- 23:00:03
I can't believe the amount of otherwise intelligent people that bought into this uncertainty nonsense.    It is a practice that far exceeds the scope of the majority of your local cal lab's workload.    I can see where it would be useful at NIST where the margin for error is miniscule and the smallest systematic error could wreak havoc on your process.    But at Transcat or Simco? Give me a break.    I wouldn't dignify most of the garbage that comes through their doors with a statement of measurement uncertainty.     Look at the recent UL post as an example.    Someone is actually taking time out of his day to research whether or not a 5520A is adequate enough to do an 87.    Putting the capacitance measurement aside, you could use that calibrator with speaker wire and still have orders of magnitude to spare.    This whole accreditation thing has gotten way out of hand.    And these sanctioning bodies perpetuate their fiefdoms by requiring you to only use labs accredited by them, thereby ensuring a circuitous stream of steady revenues.    WAKE UP PEOPLE! The other thing that gets me is the pervasive use of Type Bs that are used to make up the majority of Uncertainty Budgets that I've seen.    These are just glorified accuracy statements.    That's why so many Scopes of Accreditation look virtually indistinguishable from one another.    You can only run a 3458A through RSS one way.    The numbers are virtually meaningless in most applications.    Overkill at best, a waste of valuable time and resources at worst.    The single largest contributor of uncertainty in any measurement is the person making it.    And these Type A uncertainties are rarely, if ever addressed.    That's why two people using the same standards can come up with vastly different results.    It's all in the technique.    Good measurements are like good porn, you know it when you see it.    You can have state of the art standards but if the guy using them only has a year of experience and is only making $15/hour, chances are you aren't getting a quality measurement.    That's like putting a $149 paint job on a Mercedes.    It just doesn't make sense.    There is no substitute for experience.    And the sooner these Cal Labs come to terms with that reality, the better off we'll be.   So while you're all busy feverishly punching in numbers on a calculator to come up with some arbitrary number that has little bearing on the measurement in question, I'll do an old school 4:1 and be on the bench.   .   .   .   .   Calibrating!
Title: Re: Measurement Uncertainty is Overrated
Post by: flew-da-coup on 01-19-2009 -- 04:51:40
You are really poking a stick at a hornets nest now. :-o
Title: Re: Measurement Uncertainty is Overrated
Post by: Hawaii596 on 01-19-2009 -- 08:05:03
Bzzzzzzzzzzz!!!!! (sound of bees in the hornet's nest getting riled).
Title: Re: Measurement Uncertainty is Overrated
Post by: fwisdo on 01-19-2009 -- 12:21:59
The technicians should be on the bench calibrating.  The Metrologist or Calibration Engineer should do the uncertainty budgets. 
One problem is that as equipment gets more accurate it is harder to maintain a 4:1 all the way to NIST.  By stating known uncertainties NIST's standards do not have to be >64X more accurate than the equipment in the field.
I do, however, feel that measurement uncertainty can be way over done and waste a lot of time.  But it is the direction metrology is moving in so we will all have to deal with it.  Don't even look at Z540.3 you will really flip out. 

Process Measurement (1)
Working Metrology Laboratory (4X)
Reference Metrology Laboratory (16X)
National Laboratories. . . NIST (64X)
BIPM (256X)

Duckbutta. . . Your avatar rocks!!!
Title: Re: Measurement Uncertainty is Overrated - No its Not.
Post by: Hawaii596 on 01-19-2009 -- 14:43:58
Metrology has truly been cheapened by the Save-A-Buck mentality of too many these days.  It's unfortunate that so many less-than-reepectable labs have learned how to make uncertainty budgets look good.

But it is those same shoddy labs that created this need.  Oversimplifying the task of calibration in this technologically rapidly changing world (with such things is thinking uncertainty is nonsense) will likely one day cause the next "Challenger Disaster."

Let's take the example of the Fluke 5520A capacitance function.  If labs did good work all the time, every time with properly trained metrologists, proper procedures, proper standards, etc., then there would be no need for ISO17025.

Some use poor practices through ignorance; others through selfish motives.  Is the Fluke 5520A the correct standard to cal capacitance on a Fluke 87 DMM?  What is the cumulative uncertainty?  Is it 4:1 or better?

I don't quite understand the ignorance of belittling a scientifically valid analysis simply because you don't understand it.  If you don't know how to calculate expanded uncertainty to what ever [K] factor confidence level suits you, take a class, or pursue a different profession.  But don't cast dispersion upon something just because you don't understand it.
Title: Re: Measurement Uncertainty is Overrated
Post by: Duckbutta on 01-19-2009 -- 15:05:52
Dude,

You still don't get it. Even a cursory side by side ACCURACY comparison (forget uncertainty for a minute) between the 5520A and the 87 would have told you that you weren't making a 4:1 measurement. I don't need a calculator to figure that out. So while you're in the conference room, sipping tea with your pinkie in the air, talking with a fake British accent about Uncertain-teeee, I'll be on the bench maintaining a 4:1 ACCURACY ratio, employing good measurement practices, and KNOW that I'm making quality measurements. Meet you at the outgoing shelf....if you ever get there.
Title: Re: Measurement Uncertainty is Overrated
Post by: flew-da-coup on 01-19-2009 -- 16:05:19
There are some good reasons for having measurement uncertainties. For example, you can have 2 standard resistors both spec'd @ 1%. I can have one cal'd with a 4:1 TAR and one cal'd with actual values with Measurement Uncertainties. Now I have a 1% resistor and another that is 5ppm. The same goes with the 5520A on capacitance. If I have my 5520a calibrated with measurement uncerts. for capacitance with data then my 5520a will have a higher accuracy when doing capacitance. I think the problem here is that many in our field do not understand how measurement uncertainty works. Calculating uncerts is one thing, knowing how to use them is another. It's not about having scopes making your lab look good. It is about having higher accuracy and measurement confidence. This does not effect most field or working labs like it does primary and secondary labs.

I am sure I will be blasted for my post, but I will be more than glad to explain myself if need be. :-D
Title: Re: Measurement Uncertainty is Overrated
Post by: Duckbutta on 01-19-2009 -- 17:23:59
Coup, I'm with you on that. Uncertainties definitely have their place. I never claimed that they didn't. I just said that their use in your local cal lab is usually, though not always, overkill. To utter the words Fluke 87 and Measurement Uncertainty in the same sentence is to me just preposterous. It's just making a mountain out of a mole hole and is a complete waste of time and resources. The Type B uncertainties are negligible in that case (excluding capacitance). Hell, you'd be hard pressed to find a Type A uncertainty that would have any noticeable effect on the calibration. What's a few uV amongst friends? A simple 4:1 TAR is MORE than sufficient in that instance. And I think that is basically the point you were trying to make, Coup, not to overthink this stuff. Face it, most of the work done in your local commercial cal lab isn't rocket science and no amount of uncertainty analysis is going to change that fact. Metrologists are analytical by nature and most have a healthy dose of ego (two good traits to have in this business). That's why I think so many are susceptible to falling into the Uncertainty Trap. They view it as a chance to show everybody that they're the smartest guy in the room but just wind up sounding foolish. Some advice to Hawaii596 and those like him, go back to your bush league cal lab, bang out your little handhelds and leave the uncertainty analysis to us, the big boys.
Title: Re: Measurement Uncertainty is Overrated
Post by: flew-da-coup on 01-19-2009 -- 18:03:33
Quote from: Duckbutta on 01-19-2009 -- 17:23:59
Coup, I'm with you on that. Uncertainties definitely have their place. I never claimed that they didn't. I just said that their use in your local cal lab is usually, though not always, overkill. To utter the words Fluke 87 and Measurement Uncertainty in the same sentence is to me just preposterous. It's just making a mountain out of a mole hole and is a complete waste of time and resources. The Type B uncertainties are negligible in that case (excluding capacitance). Hell, you'd be hard pressed to find a Type A uncertainty that would have any noticeable effect on the calibration. What's a few uV amongst friends? A simple 4:1 TAR is MORE than sufficient in that instance. And I think that is basically the point you were trying to make, Coup, not to overthink this stuff. Face it, most of the work done in your local commercial cal lab isn't rocket science and no amount of uncertainty analysis is going to change that fact. Metrologists are analytical by nature and most have a healthy dose of ego (two good traits to have in this business). That's why I think so many are susceptible to falling into the Uncertainty Trap. They view it as a chance to show everybody that they're the smartest guy in the room but just wind up sounding foolish. Some advice to Hawaii596, and those like him, go back to bangin' out your little handhelds and leave the uncertainty analysis to us, the big boys.

Thanks, that is close to what I am saying. I think that when it comes to MUA there are too many pokers in the fire. This is why we commonly hear different interpretations and understandings of MUA. The MUA Bible belongs to NCSLI and it is consistently evolving. I try not to give any validity to all those other MUA publications. I think all they do is muddy the water.

   I also agree that most of the time MUA will not be needed in the working lab. However, I am sure at times it has to be implemented for specific requirements at times.

This discussion is a real good one for us to have here. I think everyone should contribute their 2 cents on MUA. It will help us have a more clear understanding and help those who are in the dark see what it is all about. I cannot say that I am an expert, I just know what I have used and still use concerning MUA.
Title: Re: Measurement Uncertainty is Overrated
Post by: MIRCS on 01-20-2009 -- 06:17:20
Quote from: Duckbutta on 01-19-2009 -- 17:23:59
The Type B uncertainties are negligible in that case (excluding capacitance). Hell, you'd be hard pressed to find a Type A uncertainty that would have any noticeable effect on the calibration. What's a few uV amongst friends?

Actually those Type B's can make all the difference in the world.  If the manufacturer would give all their data, less type B's would be needed.  Also a Type A does make a very large noticeable difference in the uncert. just by lowering the student number.

Quote from: Duckbutta on 01-19-2009 -- 17:23:59
Some advice to Hawaii596 and those like him, go back to your bush league cal lab, bang out your little handhelds and leave the uncertainty analysis to us, the big boys.

Who are you to be saying that you're the "big boys", could very well be a bush lab compared to many else out here.
Title: Re: Measurement Uncertainty is Overrated
Post by: Duckbutta on 01-20-2009 -- 08:39:09
MIRCS,

You're proving my point for me. If you read my prior posts, you'll see that I talk about the pervasive lack of Type A use in most Scopes. Type As are a critical part of ANY uncertainty analysis. A purely Type B analysis only tells half the story. The measurement is only as good as the person making it. And to omit this factor, to me, defeats the whole purpose of the analysis in the first place. Anything else is inadequate and substandard. So before you come off half-cocked, I suggest you put the calculator down, read all the pertinent threads, and then comment.
Title: Re: Measurement Uncertainty is Overrated
Post by: MIRCS on 01-20-2009 -- 09:41:36
Quote from: Duckbutta on 01-20-2009 -- 08:39:09
MIRCS,

You're proving my point for me. If you read my prior posts, you'll see that I talk about the pervasive lack of Type A use in most Scopes. Type As are a critical part of ANY uncertainty analysis. A purely Type B analysis only tells half the story. The measurement is only as good as the person making it. And to omit this factor, to me, defeats the whole purpose of the analysis in the first place. Anything else is inadequate and substandard. So before you come off half-cocked, I suggest you put the calculator down, read all the pertinent threads, and then comment.

Maybe just a little civility in the previous 15 post that you have and people wouldn't come off as half cocked.

I give a sh!t about other places scopes.  We are really talking two different things here. 

1. Uncertainity budgets

2. Scopes of companies that shouldn't be in the field at all

I personally don't like and disagree with almost every cert that has an uncertainity associated with that measurement.  I believe in the uncertainity budget over the rest.

How and or why someone couldn't take the measurements for a Type A is beyond me.  I guess I'm with the big boys as that never happens with us and we actually do budgets
Title: Re: Measurement Uncertainty is Overrated
Post by: flew-da-coup on 01-20-2009 -- 16:33:10
Where's the love? Come on guys we can discuss this in a little better way. :-D
Title: Re: Measurement Uncertainty is Overrated
Post by: Kalrock on 01-22-2009 -- 10:16:59
This argument seems kind of out of hand.   I think that most cal labs don't require every tech to be able to do Uncertainty analysis.   They have a few people that will take care of that for them.   I know that Transcat is in the process or has already complete their attempt to go ahead and have the uncertainties figured out for every standard so that their system Caltrack automatically inputs the Uncertainties.   I do think that Uncertainties are a way for the industry to set some kind of standard to try and keep people on the same page.   I mean we all are suppose to be using the same procedures it's not that different.

That being said it does seem like somebodies Uncle came up with this thing to make some money.   I mean we all know that ESD is just made up bullsh!t.   Just like we all know that resistors have magic smoke.   I still haven't managed to get high off smoking a resistor, but I try everyday.   

On the serious side if a quick check of the TAR is greater than 4:1 can't you just add 15% to your accuracies have had a solid TUR.   If I'm wrong go ahead and let me know because I'm know expert at Uncertainties.

Later
Title: Re: Measurement Uncertainty is Overrated
Post by: Duckbutta on 01-22-2009 -- 23:19:05
Kalrock,

Did you read the threads? I don't think you did because you wouldn't have trotted that tripe out there like that. That's a hangin' curve ball if I ever saw one.

That fact that Transcam is listing the uncertainties of their standards in Caltrak is the type of deception that I have been harping on. Though I don't like to admit it, I have intimate first hand knowledge of this company (Disclaimer: I was desperate and needed the money). Their uncertainties are based on the standards only, not the bush-league "techs" they hire to carry out the process. And judging by your post, Kalrock, you're one of them. Of the 10 techs at the Transcam lab where I worked, only 2 of us had formal PMEL training. The rest were your basic garden variety electronics techs. And as anyone who knows this business will tell you, that's not gonna get it. Their labs are plagued by incompotence and piss-poor management. If you are a customer seeking a legitimate cal on anything other than a handheld, I strongly suggest that you seek another provider. Transcam is in the business of selling certs, plain and simple. You can throw Simco in there too. I don't know about Davis and Sypris but I suspect it's more of the same. If you want to save yourself some money, download the accredited logo du jour, invest in a nice printer, and print your own. The ones that you print will be worth the same as the ones they print, nothing.
Title: Re: Measurement Uncertainty is Overrated
Post by: flew-da-coup on 01-23-2009 -- 05:59:13
I used to work for Sypris and they are legit. They have their ducks in a row. As for SIMCO I wouldn't work for them. I know nothing about Davis.
Title: Re: Measurement Uncertainty is Overrated
Post by: Kalrock on 01-23-2009 -- 07:37:52
I did work at Transcat in Houston and I thought they did a really good job the Uncertainties were exactly what they were suppose to be.   

Uncertainties aren't based upon the individual tech because honestly that's not a quantitative measurement and that's what calibration is mostly about.   Also as far as the Houston lab goes there was only one tech that was brought in that didn't have military cal training and he was brought in as a meter beater, but we trained him as good as your going to get.   

I think your problem is kind of like the problem I had when I worked for Davis after that.   I found it to be exactly like you were describing Transcat, but I know why it's like that and it's because of absolutely horrible management.   So don't blame your dislike for Uncertainty system because of a bad experience at a lab.   There are bad labs and good labs all over the place just like there are bad techs and good techs.   sh!t I knew techs when I was with the service that weren't worth a bucket of piss and I'm sure it's the same know matter where you go.   

So don't get mad at the system try to change it.   I sounds to me like your suggesting there should be more training and if that is something you think is pertinent then I agree.

I hope that was more on subject for you.

Thanks.
Title: Re: Measurement Uncertainty is Overrated
Post by: Duckbutta on 01-23-2009 -- 07:50:17
Coup, I'm with you on Simco. Anyone with an ounce of integrity would draw the line there. I would rather mow lawns than work for them. I'd still have my dignity and the pay would be about the same.

Putting the quality of Simco labs aside, what's the deal with their "certs"? The format is atrocious, and with a logo so archaic, shouldn't they be printing them on parchment paper? Come on, Simco! At least look the part.
Title: Re: Measurement Uncertainty is Overrated
Post by: Duckbutta on 01-23-2009 -- 08:09:45
Kalrock,

Please don't try to educate me about uncertainty analysis. I forgot more about it than you know. The process is a significant contributor of error in any measurement. Inferior technicians perform inadequate processes that are chock full of sources of error. A 10 uV uncertainty on paper can quickly become a 100 uV uncertainty in reality. And anyone that doesn't consider the process in their analysis is either lazy or incompotent. Probably both.
Title: Re: Measurement Uncertainty is Overrated
Post by: Kalrock on 01-23-2009 -- 08:36:33
Well then how is anyone suppose to know what kind of uncertainty they have?  I'm just saying if you think that there is a problem what is your solution because otherwise it just sounds like complaining.   Unless you want some answers from the community.   In that case the only thing that comes immediately to mind is standardizing the industry and that could mean everyone being required to have an ASQ cert.   I mean hundreds of other industries require some kind of standardized test to check for competence so that could be ours.
Title: Re: Measurement Uncertainty is Overrated
Post by: Duckbutta on 01-23-2009 -- 08:45:10
Google "Type A Uncertainties" and then get back to me.
Title: Re: Measurement Uncertainty is Overrated
Post by: Kalrock on 01-23-2009 -- 09:24:48
Look I know how to do Uncertainties, but that still doesn't factor out the human error element that you keep talking about.   I mean Uncertainties are really just statistics and statistics can be made to say whatever you want.   I thought that was one of the reason behind your post.   
Title: Re: Measurement Uncertainty is Overrated
Post by: Wilk on 01-24-2009 -- 15:12:45
Quote from: Kalrock link=topic=1183. msg12199#msg12199 date=1232724288
Look I know how to do Uncertainties, but that still doesn't factor out the human error element that you keep talking about.    I mean Uncertainties are really just statistics and statistics can be made to say whatever you want.    I thought that was one of the reason behind your post.    

Kalrock,
I was actually going to try help you out untill I saw that post.   It pretty much proves you don't understand uncertainties, or the reason's we use them.

Human error will alwalys be a factor of measurement uncertainty.   It will always be a portion of all type A testing.   And it will always be tested in PT's which are a requirement of all 17025 labs.   Also when you get away from calibrating 77's with 5700's the human error can be one of the largest contributions to uncertainty.

As far as statistics goes,  try making them say whatever you want, and getting them by a 17025 assessor.   See the magic behind real statistics is that they are only as good as the data that goes into them, and the guy doing the math.    Crap in, crap out theory.   With real data, I don't care what statistical meathod you chose to use, you will come up with answers that are very close to each other, or you did the math wrong.   Trust me that there are no real statistical meathods out there that will give you answers way out in left field if they are done correctly.

As far as different companies go, compare the scopes.   They are all pretty much on par.   BMC's, and qaulity of work are 2 quite different things though.   And remember contract review is a 50/50 responsibility between the vendor, and client.   Did you ask for, and pay for what you wanted.   Our did you just blindly send something in for cal? 
Title: Re: Measurement Uncertainty is Overrated
Post by: flew-da-coup on 01-26-2009 -- 04:46:59
Quote from: Wilk on 01-24-2009 -- 15:12:45
Quote from: Kalrock link=topic=1183. msg12199#msg12199 date=1232724288
Look I know how to do Uncertainties, but that still doesn't factor out the human error element that you keep talking about.    I mean Uncertainties are really just statistics and statistics can be made to say whatever you want.    I thought that was one of the reason behind your post.    

Kalrock,
I was actually going to try help you out untill I saw that post.   It pretty much proves you don't understand uncertainties, or the reason's we use them.

Human error will alwalys be a factor of measurement uncertainty.   It will always be a portion of all type A testing.   And it will always be tested in PT's which are a requirement of all 17025 labs.   Also when you get away from calibrating 77's with 5700's the human error can be one of the largest contributions to uncertainty.

As far as statistics goes,  try making them say whatever you want, and getting them by a 17025 assessor.   See the magic behind real statistics is that they are only as good as the data that goes into them, and the guy doing the math.    Crap in, crap out theory.   With real data, I don't care what statistical meathod you chose to use, you will come up with answers that are very close to each other, or you did the math wrong.   Trust me that there are no real statistical meathods out there that will give you answers way out in left field if they are done correctly.

As far as different companies go, compare the scopes.   They are all pretty much on par.   BMC's, and qaulity of work are 2 quite different things though.   And remember contract review is a 50/50 responsibility between the vendor, and client.   Did you ask for, and pay for what you wanted.   Our did you just blindly send something in for cal? 

Spot on.
Title: Re: Measurement Uncertainty is Overrated
Post by: Kalrock on 01-26-2009 -- 08:40:16
Quote from: Wilk on 01-24-2009 -- 15:12:45

Kalrock,
I was actually going to try help you out untill I saw that post.   It pretty much proves you don't understand uncertainties, or the reason's we use them.

We use TURs instead of TAR because they provide us and our customers with a more reliable and accurate understanding of our readings.  This isn't just for them, but for us as well because if you don't know that your standard and is any better than your UUT then your practically just doing a lick stick job.  Some customers like some I&E guys don't give a rats ass about TURs, but some like engineers need to know with confidence that there reading are spot on.  If you think different or have a better way of saying it or whatever go ahead and let me know.  I'm only here to learn.

Quote from: Wilk on 01-24-2009 -- 15:12:45

Human error will alwalys be a factor of measurement uncertainty.   It will always be a portion of all type A testing.   And it will always be tested in PT's which are a requirement of all 17025 labs.   Also when you get away from calibrating 77's with 5700's the human error can be one of the largest contributions to uncertainty.

I do understand uncertainties and why they are used.  I was trying to understand duckbutta's point.  I felt that his real gripe was with an individual lab that was run poorly.

Quote from: Wilk on 01-24-2009 -- 15:12:45
As far as statistics goes,  try making them say whatever you want, and getting them by a 17025 assessor.   See the magic behind real statistics is that they are only as good as the data that goes into them, and the guy doing the math.    Crap in, crap out theory.   With real data, I don't care what statistical meathod you chose to use, you will come up with answers that are very close to each other, or you did the math wrong.   Trust me that there are no real statistical meathods out there that will give you answers way out in left field if they are done correctly.

As far as different companies go, compare the scopes.   They are all pretty much on par.   BMC's, and qaulity of work are 2 quite different things though.   And remember contract review is a 50/50 responsibility between the vendor, and client.   Did you ask for, and pay for what you wanted.   Our did you just blindly send something in for cal? 

I think your right I know that you can change your k factor to give the appearance of a better TUR and I don't think it's technically wrong as long as you state it, but like you said it's a 50/50 responsibility.  Some customers want a sticker and some want their equipment calibrated.

Anyways I need to state that I don't have any problem with Uncertainties and their uses.  I was just trying to understand Duckbutta's problem.  If I'm wrong about something that's fine nobody knows everything and I sure as hell wouldn't claim to.

Title: Re: Measurement Uncertainty is Overrated
Post by: Wilk on 01-26-2009 -- 16:32:14
flew-da-coup,
Thanks, that was really the best way I could find to sum that up.

Kalrock,

If you where just playing Devils Advocate with Duckbutta then cool.   I wasn't trying to start some childish flame war or anything.   It just seemed like the conversation was getting a little off into left field from the truth.   If you are looking to learn, google is free and all the information is there.   Most all ISO documents have "free look alikes" out there that have the info needed to learn about this.   Past that look for document's, not opinions.   Opinions are everywhere, and most of them in this industry are biased for one reason or another.   I try to stay unbiased, but its tough so I just stay away from aurguments that aren't supported by documentation.

NIST, NCSLI, and ILAC's websites probably are some of the best sources for free info IMHO.   Specifically ILAC as they are the ones who set the rules for the 17025 accrediting bodies. 




Title: Re: Measurement Uncertainty is Overrated
Post by: skolito on 02-20-2009 -- 08:32:16
Just out of morbid curiosity how many of your customers actually use your readings and uncertainties in their process?????

I would say 5% if that. Most of the calibrations I do are for ISO 9000 requirements and the customer doesn't care what the reading is they want to know was it in or out, but we are a working lab not a standards lab. A standards lab is another thing in itself and I want my uncertainties from my standards spelled out in front of me when I calculate mine. as far as a Fluke 87 the 5500/5520 cannot do the capacitance at a tur of 4:1 so just use standard caps that's what we do. Uncertainties have their place and we use them as such but if a customer sends a 1.5" 300psi 5% pressure gauge that cost him $15 and he wants it done 17025 with uncertainties what is the freaking point other than $100-$150 bucks for my company its useless for them because it on a production line and isn't even looked at until I cal it the next year. Its their auditor or my biggest headache "UL" causing him to have it done.
Title: Re: Measurement Uncertainty is Overrated
Post by: Wilk on 02-21-2009 -- 12:00:26
That is going to vary greatly from lab to lab, and customer to customer.   Uncertainties on 4% devices doesn't do much other than help establish traceability.   On better equipment though proper use of uncertainties can help you realize that certain equipment can be used at accuracy levels well within the stated specifcations of a device.   Most reference standards are a great example of this.   A nice reference shunt for example could carry a specfication of 0. 1%.   When measured at the right lab, and by using the appropriate measured values and uncertainties you can easily make measurements well within 0. 05%.   

Your calibration of a 77 with a 5520 is a great example of this also.   If your 5520 shows that every time you use it, or have it calibrated that capacitence value is repatable within half of its spec, and the calibration lab you use has decent measurement uncertainies than you very well could be 4:1.   With stock specs you are about 3. 5:1, so if the 5520 is testing within 75% of the spec, you could be good to go.   This is the value add of uncertaines to a calbration lab.   Correct use of uncertainty information can prove that your standards can be used dramatically more accurate levels that the maximum allowable specifcations provided by the MFR.   It can also prove that the MFR specifcations are way to tight for that instrument unfortunately.         

Proper use of uncertainties is becoming more, and more nessecary to meet the calibration requirements of end user equipment because the manufactors are now putting what used to be metrology grade specs in the end users hands.

I can't really put a % number on the equipment that clients are using uncertainties, but I would bet big money it is way above 5%.   Alot of calibrations produce correction factors as data versus in or out of tolerance.   Accelerometers, microphones, load cells, pressure trancsducers, and so on.   Its a pretty big list.   Anyways, all of those types of equipment must be used with calibration factors, and thier uncertainties to be utilized correctly.   



   
Title: Re: Measurement Uncertainty is Overrated
Post by: CalLabSolutions on 02-22-2009 -- 12:19:47
The industry is changing...

We can no longer live by 4 to 1..  We are quickly approaching an accuracy floor, because field test instruments are becoming more and more accurate. 

We just did a procedure for a Yokagawa WT3000 Power Analyzer.  Our customer wanted to test this unit (as best they can) using the standards they had in their lab (5720, 5520, 5790 ext.).  We had to drop some test points like 1A at 1MHz and we were only able to get 2 to 1 in some points. 

We also did a procedure for the Agilent N9340A\B Spectrum Analyzer.  The Scale Fidelity test on this handheld unit rivals the ESA's specifications.  We had to test this just like we the PSA's Scale Fidelity.

I think these instruments are a good indication of the future.  Field test equipment will be almost as accurate as our lab standards. And all we will be able to do is test it as best we can, state our uncertainties and move on..

Mike....