Author Topic: 33K5-4-504-1  (Read 5012 times)

Offline MIRCS

  • 7-level
  • ****
  • Posts: 535
  • Action Taken: +9/-3
    • Email
33K5-4-504-1
« on: 08-02-2004 -- 22:00:06 »
33K5-4-504-1


Table 2

Add 2500S as the primary standard for the calibration. Accuracy 0.06˚C

3.1 through 3.4 Remain the same.

4.1.1 Connect all Sensors to the TI through the remote cables.

4.1.2 Place the TI Sensors into the Humidity Generator

4.1.3 Set the Humidity Generator for 50% RH and 35˚C and Press RUN, allow generator to stabilize at these temperatures for a minimum of 30 minutes.

4.1.4 Switch to a different sensor and take readings for each remaining sensor.

4.1.5 Repeat steps 4.1.3 and 4.1.4 with the temperature set to 60˚C.

Leave Tables 2 and 3 the same.

Make the current procedure an alternate method.

 

DISCLAIMER:  This site is not an official US Air Force site, it is intended for private use only.  It is not endorsed, in any way, by the US Air Force, the DoD, or any other governmental agency.  Additionally, all information found within this site is just that, it is NOT meant to be used in place of authorized publications.